• Marconi Distortion MeterTF2337

    From G8MNY@21:5/101 to TECH on Wed Jul 24 03:10:53 2019
    R:190724/0601Z 6126@GB7YEW.#79.GBR.EURO LinBPQ6.0.18
    R:190724/0600Z 24589@N9PMO.#SEWI.WI.USA.NOAM BPQ6.0.18
    R:190724/0556Z @:JE7YGF.03.JNET7.JPN.AS #:21621 [Ichinoseki] $:13301_GB7CIP R:190724/0554Z @:GB7CIP.#32.GBR.EURO #:13301 [Caterham Surrey GBR]

    From: G8MNY@GB7CIP.#32.GBR.EURO
    To : TECH@WW

    By G8MNY (Correction Sep 16)
    (8 Bit ASCII graphics use code page 437 or 850, Terminal Font)

    Level Meter Distortion Meter
    ┌──────────────────────────────┐
    │--┌───────────┐┌───────────┐--│ I have been working on one of these
    │ │· ││· │ │ old AF automatic distortion meters.
    │ │ '. ││ '. │ │ It is quite different to the usual │--└───────────┘└───────────┘--│ fiddly variable Wien Bridge deep
    │ │ notch type. So it is ideally suited
    │ 0 │ for setting up tape machines, where
    │ Function Distort │ the playback level & frequency will
    │ [][][][] [][][][] │ never be that steady.
    │ O o │
    │ [][][][][][][][] │
    │ O ø │
    └──────────────────────────────┘
    In Level Range Mains

    Level Ranges:- 30V, 10V, 3V, 1V, 300mV, 100mV, 30mV, 10mV.
    :- 30, 20, 10, 0, -10, -20, -30, -40 dBV


    Response :- 10Hz - 250kHz @ -3dB

    Distortion Filters 400Hz, 1kHz & External.

    Distortion Ranges: 30%, 10%, 3%, 1% min to 0.2%
    & dBs -10 -20 -40 -40 min to -55dB

    Harmonic Response: 1.5x Filter to 130kHz @ -3dB

    Mains 230V. odd 3 pin 1cm square socket.

    The automatic refers to the distortion measurement does not need the zero level reference level to be set first. (But is must be reading 30-100% on the level meter scale.) This is done not by a very low distortion AGC system as you might think, but a cleaver meter movement with 2 coils & no return springs. One coil is fed with rectified DC from the input level detector 1 & acts as the return spring & the other is fed from an 2nd detector from a high pass filter that removes all of the fundamental, so it just the harmonics. The resultant meter deflection is the distortion fraction.

    In ┌─────┐ ┌─────┐ AC┌─────┐ ┌───────────┐ o──┤Level├─┤Input├─┬────┤Level├──────────────┤LEVEL METER├─────┐
    │Atten│ │ Amp │ │ │Det 1│ DC └───────────┘ │
    └─────┘ └─────┘ │ │ ├──────────────────────────────┐ │
    │ └─────┘ ┌────┴─┴─────┐
    │ ┌─────────┐ │ DISTORTION │
    │ o─┤400Hz HPF├─o ┌─────┐ ┌─────┐ │2 COIL METER│
    │ ├─────────┤ │ Set │ │Level│ └────┬─┬─────┘
    └>o─┤ 1kHz HPF├─o<─┤Gain ├──┤Det 2├───────┘ │
    └─────────┘ │ Amp │ │ ├─────────┘
    o─ External ──o └┬┬┬┬┬┘ └─────┘ DC
    ^──┘Dist Range

    It just uses simple RC transistor Amps & bridge diodes for the electronics. The notches are sealed LC high Q units. (M derived?)

    There are only 3 presets:- Level gain, Distortion gain, & input amp Bias.

    DISTORTION MEASURING PRINCIPLES
    f1 f2
    │ f1 0dB─┤ | ┌┴────── │ f1
    Level│ │ f2 │ | │ High │ f2
    │ │ │ f3 │ | │ Pass │ │ f3
    │ │ │ │ f4 f5 -60dB_│___|_/ Filter │ │ │ f4 f5
    └───┴───┴───┴───┴───┴ └────────────── └───────┴───┴───┴───┴
    Distorted Signal Filter Filtered Signal

    The high pass filter must completely remove the fundamental & all LF hums etc. but leave all the harmonics intact. The detectors (especially the distortion one) should be a true RMS type, as the waveform will generally be very complex & nothing like a sine wave. The ratio of f1 to all the harmonics f2+++ is the distortion factor.

    Calibration is done by adding in a 2nd tone (e.g. 3-10x freq) via high value Rs at say 10% of the level to indicate 10% distortion. The distortion meter bandwidth can also be checked with this circuit.

    1V 1kHz ─o-o───10k──┬───┬──────┐
    Osc 1 o │ │ │
    _│_ │ │ Distortion
    │ DVM Meter
    100mV │ _│_ _│_
    2-20kHz ─o-o───10k──┘
    Osc 2 o
    _│_

    To check the HPF filter is working OK a very pure sine wave is needed, so an LC tuned circuit after a signal generator is used. But I did find that > 0.5V @ resonance on the meter did make the inductor distortion rise > 0.01% !

    Osc 88mH 500mV
    50mV 1kHz ───((((()──┬─────┐ Hi Z
    low Z │ Distortion
    │ 0.28uF === Meter
    _│________________│_____│

    MODERN ANSWER
    Of course with a good quality PC sound card & a Furrier analysis program that can add up the harmonics distortion measurement should be easy.


    See my bul "Simple 1kHz AF osc".


    Why don't U send an interesting bul?

    73 de John, G8MNY @ GB7CIP